      how can I adjust the correct temperature for heating bed?
      if I set 60 and measure with ir thermometer then bed has only 45 degrees

      1 Reply Last reply Reply Quote 0
      • dc42undefined
        dc42 administrators
        last edited by

        If the bed is glass, it's normal for the temperature on the top to be 5C to 10C lower than the bottom where the thermistor is.

        Have you set the correct B value for the bed thermistor you are using, in the M305 P0 command?

        Duet WiFi hardware designer and firmware engineer
        Please do not ask me for Duet support via PM or email, use the forum
        http://www.escher3d.com, https://miscsolutions.wordpress.com

        1 Reply Last reply Reply Quote 0
        • Altaezundefined
          Altaez
          last edited by

          M305 P0 R4700 H0 L0

          but I do not know if it is right

          1 Reply Last reply Reply Quote 0
          • Altaezundefined
            Altaez
            last edited by

            I have now resisted 100k

            1 Reply Last reply Reply Quote 0
            • dc42undefined
              dc42 administrators
              last edited by

              There is no B parameter in your M305 P0 command, so the firmware is using the default value of 3988. Do you have any information about the bed thermistor in your printer?

              Duet WiFi hardware designer and firmware engineer
              Please do not ask me for Duet support via PM or email, use the forum
              http://www.escher3d.com, https://miscsolutions.wordpress.com

              1 Reply Last reply Reply Quote 0
              • Altaezundefined
                Altaez
                last edited by

                I have silicone mat, aluplatte, glass
                M305 P0 T100000 B3988 R4700 H0 L0
                is it right?

                1 Reply Last reply Reply Quote 0
                • Altaezundefined
                  Altaez
                  last edited by

                  resistance of thermistor has 103k

                  1 Reply Last reply Reply Quote 0
                  • Dougal1957undefined
                    Dougal1957
                    last edited by

                    if it is the std Chinese type Silicon mat then they usually have a thermistor with a B Value of 3950 but this is at temps of 25/50 I used DC42's calculator and it gave a B value of 4008 for the typical temps of 100 degrees.

                    1 Reply Last reply Reply Quote 0
                    • Altaezundefined
                      Altaez
                      last edited by

                      yes, silicone mat is from china
                      M305 P0 T100000 B4008 R4700 H0 L0

                      it is OK?

                      1 Reply Last reply Reply Quote 0
                      • dc42undefined
                        dc42 administrators
                        last edited by

                        Yes. However, increasing the beta value in M305 will reduce the temperature reading, for temperatures above 25C. So try B3950.

                        Duet WiFi hardware designer and firmware engineer
                        Please do not ask me for Duet support via PM or email, use the forum
                        http://www.escher3d.com, https://miscsolutions.wordpress.com

                        1 Reply Last reply Reply Quote 0
                        • Altaezundefined
                          Altaez
                          last edited by

                          Now I have set 70 degrees and have 62.
                          can one adjust more exactly?

                          1 Reply Last reply Reply Quote 0
                          • deckingmanundefined
                            deckingman
                            last edited by

                            Is that 62 measured with your IR thermometer? If so, then how accurate is your IR thermometer? They can often be quite a way out.

                            Ian
                            https://somei3deas.wordpress.com/
                            https://www.youtube.com/@deckingman

                            1 Reply Last reply Reply Quote 0
                            • Dougal1957undefined
                              Dougal1957
                              last edited by

                              also bare in mind that with these Silicon heaters the thermistor is usually on the underside of it and is below the bed itself so you will get a difference anyhow but you can adjust the B Value to get it closer to what you want but As Ian (Deckingman) has pointed out do you know your IR Thermometer is accurate I would use a multi meter with a Thermocouple probe that is in very close contact with the top of the glass

                              1 Reply Last reply Reply Quote 0